bud-bennettBud Bennett 07/23/2020 at 23:132 Comments
It's been more than 1 year since I purchased the Keysight 34461A DMM. It is officially out of calibration now. Time to do a quick measurement. I powered up the voltage calibration box (the Digikey box) and the DMM for six hours prior to taking a measurement. Here's the results:
DIgikey Box:
Ambient Temp: 68°F(20C)
DMM settings: 100PLC, Auto Input Z (High Z)
Reading (not averaged)    Scale
99.9994mV (nulled)          100mV
1.000011V                          1V
2.49976V                          10V
4.99940V                          10V
10.00003V                         10V
These are not very far off from the six month readings.
AliExpress Box:
(same settings and ambient temp as above, soak time 4 hours)
Reading                                            Scale
100.00181mV (avg/20 samples)        100mV Nulled
1.000035V (avg/20 samples)           1V
2.50014V                                           10V
10.000114V (avg/20 samples)              10V
Current Reference(all readings 100plc, averaged with 20 samples):
Reading             Scale
10.001474mA     10mA
0.999977mA     1mA
99.99313µA       100µA
General Conclusions:
All of the voltage references seem to be very stable now. No significant difference between the USA components and the Chinese components. The current readings are drifting a bit, but the specs of the 34461A are quite a bit looser for current than voltage. The only way to tell for sure is to compare readings to a calibrated DMM with >6.5digit capability. That's not happening anytime soon.
Bharbour wrote 07/24/2020 at 03:12 point
Under a millivolt drift in a year, pretty nice!
Bud Bennett wrote 07/24/2020 at 14:20 point
Yes...either the two instruments are stable, or they are drifting in the same direction.